UPLC-MS QTOF pressure drop every 15 seconds

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hello,

I am having a problem with pressure and retention time with Waters UPLC QTOF system.

Retention times are about 1-1.5 min late.

Pressure is at ~11000 psi (3000 psi higher than normal) and drops about 3000 psi every ~20 sec interval consistently. I've already ruled out the problem stemming from the LC, since this pressure issue does not exist when the MS is turned off and only the LC is on. Once the MS is on and flow is going through the sample probe, I begin to see this pressure issue.

Any advice on how to correct?
If it is just whether the MS is powered on or not, I would look at the position of the divert valve. Is it switching or partially switching and causing a blockage in the flow?

Or possibly the heating of the needle causing a change in flow. Not many things in the MS are involved in the flow that would cause restriction.
The past is there to guide us into the future, not to dwell in.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry