Hi there, thanks in advance for any help or advice!

I'm using GC MS for quantitation. Originally I used selective ion monitoring mode to quantify analytes of interest, which I have commerical standards of. So I have been able to make a calibration curve for each and quantify those using specific ions.

However, I would now like to look at all the compounds in the samples and see % composition and change. But I don't have standards for each, so I can't make calibration curves.

I ran the GC MS in full scan mode and SIM mode so I did try using the TIC to look at each compound as a percentage of the total area. However, a lot of the compounds don't separate well and peaks merge, so I can't really use the TIC.

Is there any way to use the extracted ion chromatograms of analytes and get a % composition without calibration curves of each? I did add an internal standard to each sample as well.

Thank you, I hope this made sense!
Not without standards. TICs is short for tentatively identified compounds and they're called that for a reason.
Not without standards. TICs is short for tentatively identified compounds and they're called that for a reason.
???

I've never known TIC to be anything other than "Total ion chromatogram" or some variant thereof-basically the chromatogram of everything hitting the detector. I've done quantitation based on TIC areas(whether with a calibration curve or IS) for as long as I've been doing GC-MS.

To the OP:I'd want to see your full method, but there should be a valid way to quantitate based on an extracted chromatogram provided that you build a calibration curve based on it and/or an IS. I have done this before to tease out difficult to separate compounds, although generally I was only quantifying one of them. Of course this assumes the co-eluting compounds each have a signature ion that you can use.
Not without standards. TICs is short for tentatively identified compounds and they're called that for a reason.
???

I've never known TIC to be anything other than "Total ion chromatogram" or some variant thereof-basically the chromatogram of everything hitting the detector. I've done quantitation based on TIC areas(whether with a calibration curve or IS) for as long as I've been doing GC-MS.

To the OP:I'd want to see your full method, but there should be a valid way to quantitate based on an extracted chromatogram provided that you build a calibration curve based on it and/or an IS. I have done this before to tease out difficult to separate compounds, although generally I was only quantifying one of them. Of course this assumes the co-eluting compounds each have a signature ion that you can use.

Hi, yes thank you, by TIC I meant total ion chromatogram.

I'm looking at sterols and I have a calibration curve for one sterol, the most abundant in my data, made using a standard (and an internal standard). But there are other structurally similar sterols (isomers or one double bond different) which I don't have standards for.

And those which co elute do have different identifying ions to each other, so I can get nice extracted ion chromatograms. I worked out some values with the TIC and internal standard, like with an FID, but I think it's misleading due to the coelution.

I can't collect more data on the same instrument so I'm hoping I can figure out a way with what I have.

For the isomers, which have the same major ions as the sterol I originally quantified, could I use the same calibration curve to quantify using the EIC? But then what about the ones with different ions?

Thank you!
4 posts Page 1 of 1

Who is online

In total there are 20 users online :: 1 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Ahrefs [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry