Advertisement

LC-MS of Nucleotides.

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hello!

My team and I are attempting to use a Thermo TSQ Quantum Ultra triple quad for the detection of nucleotides. However, their TIC and subsequent mass spec looks awful (below).

Image

They're using 10 mM ammonium acetate and methanol as their LC eluent. I have tried optimizing the MS a few times but have had a hard time getting better ionization, so I was curious if there are any tips someone could provide?

They're using positive ion mode and I've had them try adding formic acid to their samples but nothing really helped.

I'm happy to answer questions as needed.

Thanks!
J Am Soc Mass Spectrom. 2019 Jun; 30(6): 987–1000.
Published online 2019 Mar 7. doi: 10.1007/s13361-019-02140-7
PMCID: PMC6520184
NIHMSID: NIHMS1523504
PMID: 30847833
Simultaneous Quantification of Nucleosides and Nucleotides from Biological Samples
Liqing He,1,2,3,4 Xiaoli Wei,1,2,3,4 Xipeng Ma,1,2,3,4 Xinmin Yin,1,2,3,4 Ming Song,2,3,5 Howard Donninger,5,6 Kavitha Yaddanapudi,5,6 Craig J. McClain,2,3,5,7,8 and Xiang Zhang1,2,3,4,7

Have you seen this paper from 2019??
It is available on-line, if you do not have access to JASMS.
Has a LOT of experimental detail!!

Regards,
JMB
J Am Soc Mass Spectrom. 2019 Jun; 30(6): 987–1000.
Published online 2019 Mar 7. doi: 10.1007/s13361-019-02140-7
PMCID: PMC6520184
NIHMSID: NIHMS1523504
PMID: 30847833
Simultaneous Quantification of Nucleosides and Nucleotides from Biological Samples
Liqing He,1,2,3,4 Xiaoli Wei,1,2,3,4 Xipeng Ma,1,2,3,4 Xinmin Yin,1,2,3,4 Ming Song,2,3,5 Howard Donninger,5,6 Kavitha Yaddanapudi,5,6 Craig J. McClain,2,3,5,7,8 and Xiang Zhang1,2,3,4,7

Have you seen this paper from 2019??
It is available on-line, if you do not have access to JASMS.
Has a LOT of experimental detail!!

Regards,
JMB

I have not but greatly appreciate the share and will give this a shot and post back with the results!
3 posts Page 1 of 1

Who is online

In total there are 7 users online :: 1 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Ahrefs [Bot] and 6 guests