During analysis, we are observing droplets of liquid forming on the tip of our ESI probe.

Is this related to the desolvation temperature or gas flow? How can we avoid this droplet formation.
Hi
Liquid droplets on the tip of the probe will be due to insufficient gas flow to produce a spray, or an excessive liquid flow (blocked splitter ?).
Droplets on the windows of the spray chamber will be due to insufficient gas temperature causing condensation.
This could be caused by:
too much mobile phase flow or
insufficient nebulizer gas flow/pressure (whichever setting you have)
insufficient desolvation gas flow/temperature
(I have never observed liquid droplets on the observation window - maybe insufficient desolvation gas flow/temperature or a poorly performing source waste/vacuum flow system)

Mass spectrometer manufacturers usually provide guidance table on settings for various mobile phase flow rates and compositions. (It is important to pay attention to the minimum values for your mobile phase flow rate.)
3 posts Page 1 of 1

Who is online

In total there are 9 users online :: 1 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Bing [Bot] and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry