Calibration Valve 5973 always open?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Dear fellow chromatographers,

our MS System (Agilent 5973) shows steady peaks of the calibrant PFTBA during all acquisitions. If the valve is switched off under manual tuning the ion peaks do not disappear and change in abundance. Also if the valve is getting purged there is no click sound. Did anybody experienced something like this? Could it be that the valve is steadily open and can't manage to close?

I would be very grateful for any advice!
hi mvlo

sounds like the valve / solenoid is permanently open.
either the valve is defective or the control voltage to the solenoid is continuously on.
you could try to unplug the solenoid connector which is behind the MS lower panel. this will tell you if the valve closes when no voltage is present.
I agree. If it goes away when unplugged then there is a problem with the control board, if it stays then the valve is stuck open and probably needs to be replaced.

I have had them fail before, both open and closed.
The past is there to guide us into the future, not to dwell in.
3 posts Page 1 of 1

Who is online

In total there are 18 users online :: 1 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Ahrefs [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry