unresponsive flow diverter valve in Sciex 5500

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hi,
does anybody know how to specifically reset only the flow diverter valve module in Sciex 5000 series instruments, without performing a total hardware reset?

Sometimes, I encounter a problem that the integrated flow diverter valve stays in status "run" even if the run has stopped, and becomes unresponsive to further manual or software commands. The only way of solving this issue is to perform a hardware reset with pumps turning off, so I'd be really grateful if someone found a way of resetting only this module in some sort of way.
Thanks and best regards,
Jurij
Hi
Did you try to deactivate and reactivate the hardware profile?
Good luck
Jörg
Hi, Jörg,
Thanks for your answer! Yes, I have deactivated the profile, and reactivated it, and unfortunately, it does not reset the valve controller. I even tried shutting down the HPLC and PC, but it does not work either. The only option is to reset or power-cycle the mass detector.
Best regards, Jurij
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Ostravaczech and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry