Hello everyone,

in a LC-MS/MS method, I am facing serious sensitivity issues with some pesticides. About 90% of the measured pesticides appear great, but the following are barely visible in 1 ug/mL in the standard solution:
Clopyralid, MCPB, 1-naphthyl acetic acid, 2,4-DB, aminopyralid, mesotrione and others

Any suggestions?
Not all pesticides work with LC and not all work with GC.

Compare this list that work well with LCMSMS
https://www.restek.com/en/products/refe ... ure/40244/

To this list that work well with GCMSMS
https://www.restek.com/en/products/refe ... ure/41952/

I notice that 2,4-D is not in either of those lists, but I have used LCMSMS to analyze for it before, but the sensitivity is less for it than most other analytes.

The ones you have trouble seeing my require different polarity of the source, or different temperature and flow rates versus the rest of your list. This can be a problem when trying to do a large list in one injection. You cannot optimize for all the analytes at the same time, so some will have good sensitivity and others not so good. May need APCI and not ESI. Not sure of your instrument setup to know what options you have.
The past is there to guide us into the future, not to dwell in.
Most of the pesticides you mentioned are acids, and would be more sensitive using negative ionization and an eluent containing acetic acid rather than formic. What was your method (instrument, LC and MS parameters)?
3 posts Page 1 of 1

Who is online

In total there are 11 users online :: 2 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Bing [Bot], Google [Bot] and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry