Hi,

does anybody know how to fix a firmware that became corrupted because of an unsuccessful update (some *.o files were not copied to the SC)?
The topic appeared in the past, apparently without any final solution.
Symptoms are well known: MS is stuck while rebooting, showing a 'Loading MSD firmware'. I can access the smartcard by telnet/ftp, but apparently, it lacks a "boostHost."

Thanks for any help

PC
If you delete the instrument in MS Configuration then reinstall it, does it go through the firmware update process? I believe that is where it usually updates it and maybe that would force it to complete a new update.
The past is there to guide us into the future, not to dwell in.
Thank you James for your advice.

an MS witout a running FW does not pass the RPC test.

I guess one issue that determined a corrupted FW may be connected with the request of USER/PASS during the update:


C:\msdchem\MSexe\firmware\5973N\msupdate.exe version 3.98
Can't get current MSSC revision. MSSC not active...
Current OS version: 1.2/22T

SC3type = SC3Plus

Update MSSC version: 5.02.07
Update OS : skipped

MSD SmartCard at 10.1.1.102 will be updated

!!>>DO NOT DISRUPT POWER TO MSD DURING THE UPDATE.<<!!

OK to proceed (Y=YES): Y
Clearing /flash/:
Found an existing BOOT.ASC file.
rm "/flash/MSSC.O"
value = 0 = 0x0
-> rm "/flash/LUI.O"
value = 0 = 0x0
-> rm "/flash/STARTUP"
value = 0 = 0x0
-> rm "/flash/httpd.o"
value = 0 = 0x0
-> rm "/flash/wdsmall_run.o"
value = 0 = 0x0
-> rm "/flash/BLINK.O"
value = 0 = 0x0
->
Sending 10.1.1.102:/flash/MSSC.O
USER and PASS required
Sending 10.1.1.102:/flash/LUI.O
USER and PASS required
Sending 10.1.1.102:/flash/STARTUP
USER and PASS required
Sending 10.1.1.102:/flash/httpd.o
USER and PASS required
Sending 10.1.1.102:/flash/wdsmall_run.o
USER and PASS required
Sending 10.1.1.102:/flash/BLINK.O
USER and PASS required
Initiating flash burn.
diskFlash
erasing bank 1: sector 0
programming data...
flash update complete...
value = 0 = 0x0
->
Code download complete.
Power cycle instrument to complete update process.
.............................................
SOLVED.
INstalled Chestatin E.02.02 and the firmware was regularly updated to 5.02.14.
The USER and PASS required remained, but the procedure completed successfully
4 posts Page 1 of 1

Who is online

In total there are 10 users online :: 0 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry