Dear all,

I've been using LC-MS for some time now, but unfortunately, I have no mentor to ask all the questions in mind so I will really appreciate your expertise.
My main concern is the hardware and keeping it clean and undamaged. There have been a few gas shortages and I would love to know the best practices for a situation of this kind.

The instrument in question is QTRAP. I've been told that if there's no gas flow I should monitor the system vacuum and if it falls below a certain level there's a risk of instrument contamination.

Why would the system vacuum keep increasing over time if there is no gas flow? The normal operating pressure of the instrument is 2x10^-5 Torr (system running) / 0.6x10^-5 Torr (system idle). I assume the decrease in vacuum when running is caused by the increased gas flow (compared to idle)?

Now, the normal (idle) pressure being 0,6x10^-5 Torr, what would be a harmfully low pressure for this system that could cause contamination?
How long, approximately, is it safe to have the instrument with full vacuum in case of gas shortage? The LC would be detached at this point, of course.

Thank you for reading and I hope you’ll find the time to share your knowledge.
Stay healthy, everyone!
In standby there is a sweep of gas between the inner and outer cones that keeps particulates and other air born contaminates out of the MS. I would suspect that if the vacuum begins to be better(lower ultimate pressure) then it would mean that the inner orifice is becoming clogged, which also happens if the samples are really dirty. I don't know that it would mean that the internals are becoming really dirty though, unless there is a lot of contaminates in the lab air.
The past is there to guide us into the future, not to dwell in.
Possibly your instrument has a gate between the high vacuum section and the splitter cones.

My PE Elan 6000 is an ICP-MS but it has analogous behaviour. When at idle under vacuum, there is a gate that actually shuts off access to the MS high vacuum section and vacuum drops to 0.9E-6. When operating, the plasma the gate opens and then the final vacuum rises to 2.5E-5.
3 posts Page 1 of 1

Who is online

In total there are 10 users online :: 0 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry