Hi all,

today i had the problem of bad fumonisin chromatography for the second time when installing a new column. The peaks of fumonisins are totally gone or can not be interpreted. The type of column we are using is Hypersil gold 10cm*2.1mm*1.9um from thermo. Mobile phases: A: 2mM NH4FA in water and B: 10mM FA in MeOH. Gradiënt from 95% A to 1%A in 8 minutes. Flow 500ul/min.

What I noticed today is that the problem seems to disappear after injection of 10-20 real matrix samples. After that fumonisin peaks are back and sharp. I presume it might have something to do with the matrix shielding off active sites (free silanols?) or irreversible retention of fumonisins until the active sites are saturated. I wonder what your thought are on this matter and if you have seem similar strange behaviour of fumonisin retention. I'd rather know the cause of the problem because it offers insight on actual mechanisms happening inside HPLC columns.

Thanks for the information!

Kind regards,

Jasper
I have approached the same problem with Biphenyl column. After reading your solution a year ago, I have used the same procedure and it worked.

But yesterday the problem have come back. I have lost all peaks for fumonisins (FB1, FB2, FB1_IS), but this time your solution didn't work. What to do now? Can you help me, please?
2 posts Page 1 of 1

Who is online

In total there are 11 users online :: 0 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry