I am developing a method for DMP,DEP,DBP,DEHP,DNOP (Phthalates) using LCMSMS at 20 ppb level. Problem is I am not able to eliminate blank interference. I tried with different eluents (0.1% formic acid and acetonitrile, methanol) and columns (C18, C8, phenyl etc). All the cases I am getting blank interference at the same transition of DMP,DBP. Can anybody suggest a solution
What is the source of your blank matrix? Often we get phthalate background in even or good RO/DI water. Usually it requires a special filter pack for low phthalate water. We had to special order them to completely remove the background levels from our water system. Almost anything made of plastic has phthalates in it so it is difficult to keep blanks clean.
The past is there to guide us into the future, not to dwell in.
Thank you James.
we tried initially with 0.1% fromic acid in water and methanol. Then with only methanol. Yesterday we did an experiment with Pure methanol and acetonitrile as diluent. Blank also Acetonitrile. A dip in interference observed. but significant, the requirement is 10 ppb
3 posts Page 1 of 1

Who is online

In total there are 4 users online :: 0 registered, 0 hidden and 4 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 4 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry