Chemstation sequence saving PDF

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hi Everyone,
I am using the set up listed below, I am having an issue that every time I am running a sequence it runs a single sample then needs me to name the pdf copy of the sample file then save it before it moves on to the next sample in the sequence, does anyone know of a way to bypass this feature ?

GC Agilent 6890N (G1530N)
MS Agilient 5973 network Mass detector G27579A
Autosampler 7683
Software
Agilient MSD Productivity Chemstation for GC and GC/MS systems. Data analysis application
(MSD Chemstation G1701EA E.02.02.1431)
Thanks,
Chemiger
99 problems and chromatography is every one !
It is because you have your printer listed as printing to PDF and it requires a unique name for each time it prints. It also means you have it automatically running Data Analysis at the end of each analysis run. You can either send it to a regular printer or turn off the Data Analysis with each run by unchecking the Data Analysis check box in the Run Method screen that pops up before you begin the sequence.
The past is there to guide us into the future, not to dwell in.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry