High split ratio alternative

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
Task:
Switch to N2 carrier gas

Current practices:
He carrier gas, backpressure-regulated GC's (Agilent). A decent amount of work is injecting neat solvents for purity, use 400:1 split ratio to prevent overload. He is limited by EPC to a total flow of 1250mL/min.

Challenge:
N2 is limited by EPC to a total flow of 200mL/min. The 400:1 split ratio isn't possible.

Any suggestions?
Dilute the sample.
Are your impurities overloading the column? If not, you could try a splitless injection to get rid of the solvent tail.

Gasman
What detector are you using?

What are the dimensions of the column you are using?

Different options can be found depending on more information.
The past is there to guide us into the future, not to dwell in.
Phosgina,

what is keeping you from using H2 instead of N2?

Best regards,

AICMM
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry