7673 injection volume error

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Yes, we are still running multiple 7673s! Out of three, one tower is currently functioning :lol:

This is actually an error I have seen before, and I am pretty sure we had a spare motor on hand to fix it - what's happening is, with a 2uL injection volume, the plunger carrier will occasionally miss the second position on the little plastic ratchet and jump to the third position, resulting in a 3uL injection. There is a small motor with a spring that controls the ratchet, but I can't seem to track down this part on the Agilent site, nor can I find a detailed schematic of the tower.

Anyone know this part number, or if there's a simple fix?
As far I remember - there is plastic wedge under the motor, it is used to adjust tension in driving belt.

Try to insert the wedge deeper, just one click.
This can help not miss the "stops" on plastic "finger" while sample loading.
It can also be the plastic "finger" with the stops that is not moving back into position fast enough. That is a simple solenoid that moves the plunger attached to the finger. If it gets dirty it can move slowly and the motor pulls the plunger past the stop before it moves back into position. Usually a little WD40 will clean the plunger and get it moving faster again.
The past is there to guide us into the future, not to dwell in.
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry