I am having an issue with my GC-MS where I have repeating peaks in my baseline. The attached images are from no-injection blank runs with a gradient from 100 to 200 C; however, this occurs under isothermal conditions as well.

The second image shows the difference at different flow rates where the time delta between peaks is very similar (~0.14 min).

The mass spectra of these peaks are showing a primary peak at 32 with a second peak at 40 (both show spikes in their respective EICs).

I've tightened all connections and "baked out" MS. The air/water check in the MS is all fine (H2O: 1.2%; O2: 1.0%; N2: 3.3%)

Has anyone encountered this before?

Image

Different flow rates (0.8 & 1.2 mL/min)
Image
Whatever it is, it is happening every ~7 seconds. That should help you track it down. I'd look for something opening/closing unexpectedly, power fluctuations, a relay going bad... If you can't find a setting somewhere or hear/see whatever it is, you may need to bring service in to check the electronics.

That it continues at the same rate despite a change in flow would make me want to focus on the MSD, not the GC.
Thanks,
DR
Image
Thanks for the reply DR. There's nothing visible/audible corresponding to the peaks. I can't say I know enough about the electronics to check it, but the fact that only select ions are producing the peaks, that led me to believe it was not electronic. I could be incorrect in this assumption though.
I'm pretty much a MSD novice myself, I'd be inclined to call the vendor for a consultation or service...
Thanks,
DR
Image
32 and 40 sound like it would be oxygen and argon causing the peaks, which would come from air. What is the scan range during the run? If scanning above 28 you could me missing the nitrogen from an air leak.

What are more of the details of the instrument? Injection port, injection type, MS and GC models, scan range. More details the better to help trouble shoot.
The past is there to guide us into the future, not to dwell in.
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry