- Posts: 68
- Joined: Sat Feb 12, 2011 1:31 pm
Solvent for NPD
Discussions about GC and other "gas phase" separation techniques.
4 posts
Page 1 of 1
I am planning to use GC with NPD for phosphorus containing compound. Which solvents are recommended to dissolve compounds? Can they contain "O", "Cl", "N"?
-
- Posts: 1742
- Joined: Fri Feb 09, 2007 10:33 pm
- Location: Albq, NM
Push,
Don't recommend CL. Do recommend turning off power for solvent elution. Really recommend you get a hold of Paul Patterson at DET (det-gc dot com)
The GURU of that detector.
Best regards,
AICMM
Don't recommend CL. Do recommend turning off power for solvent elution. Really recommend you get a hold of Paul Patterson at DET (det-gc dot com)
The GURU of that detector.
Best regards,
AICMM
-
- Posts: 3475
- Joined: Mon Jan 07, 2013 8:54 pm
- Location: Western Kentucky
Definitely don't want a nitrogen containing solvent like pyridine since the detector is sensitive to nitrogen, that would give a massive solvent peak.
I know hexane and MTBE are used when doing NP pesticides by the EPA methods though.
I know hexane and MTBE are used when doing NP pesticides by the EPA methods though.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 1108
- Joined: Wed Dec 22, 2010 10:17 pm
- Location: Chicago
For that matter you might want to avoid a GC column with nitrogen in the stationary phase like 1701, 624 (cyanopropyl side chains). Otherwise any non nitrogen solvent is probably fine.
I like
acetone (one of my favorites it dissolves almost anything)
toluene (for higher boilers also has broad solubility)
heptane
diethyl ether (I try to avoid due to risk of peroxide formation)
ethyl acetate
methanol
ethanol
isopropanol
I like
acetone (one of my favorites it dissolves almost anything)
toluene (for higher boilers also has broad solubility)
heptane
diethyl ether (I try to avoid due to risk of peroxide formation)
ethyl acetate
methanol
ethanol
isopropanol
4 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science