Inlet Temp: 300ºC
FID Temp: 300ºC
Flow: Helium at a constant 35cm/s
Oven Gradient: 200ºC for 5min > 10ºC/min to 280ºC > 280ºC 2min
Injection: 2uL injection volume with a 100:1 split ratio
Last week I performed a method validation on a method with the above parameters. I probably had 50 injections in the sequence, and everything ran great!
Yesterday I ran the method again on a storage stability test of one of our formulations, using the now validated method and I am seeing a consistent decrease in response for two of my three analytes. Every injection from the same vial will produce a lower response (this is seen in both the standard vials as well as sample vials).
I have checked the inlet, replaced the septum, gold seal, inlet liner. I've watched my injections and there doesn't seem to be anything wrong with the syringe. No leaks or pressure loss anywhere.
Literally just running the same method a couple days later, with no other method or analysis being run on the GC between then and now, and now I'm seeing this phenomenon.
Right now replicate injections are coming from the same vial (as they did in the method validation procedure) but I will probably try to inject the same solution from different vials to see if that produces any different results.
Any other suggestions would be appreciated.