Does anyone have any experience with hpmc assay?

I ran six sample preparations of a 2208 sample and got acceptable results but when I ran a 2910 sample I had high methoxy results.

The usp method is a little vague in that the injection volume can be between 1ul - 2ul. The amount of adipic acid can be between 60-100mg.

I tried both 1ul and 2ul injections and 2 appears to give me more consistent results. I read a technical document from Dow where they restrict the amount of adipic acid to 85+\-5mg (which I did do) but would the amount of adipic acid cause abnormally higher results for 1 substitution type but acceptable for another? Is there some other technique I should consider? :?:
I also noticed that the new USP method for HPMC has a 40:1 split. The old USP method used a Packed column and did not list a split ratio, assuming it was a splitless injection. The new method lists a DB-1 30m X 0.53mm, 3um column with a 40:1 split. A document I found online from DOW (before they updated the USP method) states that if you want to use an equivalent capillary method they recommended a DB-5 30m X 0.53mm, 3.0um with a 5:1 split. Besides the differences between DB1 and DB5, the USP recommendation of 40:1 split seems a little high for a 0.53mm column. Would such a high split ratio affect reproducibility of individual peaks?

It appears that my internal standard (n-octane) is lower for my samples than my standards, though I understand that is the point of an internal standard to account for injection variability.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry