-
- Posts: 8
- Joined: Thu Dec 13, 2018 6:25 pm
Most of the resources that I have found only discuss introducing liquid samples to the GC. How can I connect my Mixed Gas Instrument to the GC and control injection of samples?
Thank you!
Discussions about GC and other "gas phase" separation techniques.
James_Ball wrote:
Probably the easiest way would be to use a switching valve and a 1-2ml sample loop. Let the effluent of the permeate flow through the valve and loop then switch the loop into the carrier flow and allow the carrier to sweep a few loop volumes onto the column, then reset for next sample.
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.