Solvent for NPD

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
I am planning to use GC with NPD for phosphorus containing compound. Which solvents are recommended to dissolve compounds? Can they contain "O", "Cl", "N"?
Push,

Don't recommend CL. Do recommend turning off power for solvent elution. Really recommend you get a hold of Paul Patterson at DET (det-gc dot com)

The GURU of that detector.

Best regards,

AICMM
Definitely don't want a nitrogen containing solvent like pyridine since the detector is sensitive to nitrogen, that would give a massive solvent peak.

I know hexane and MTBE are used when doing NP pesticides by the EPA methods though.
The past is there to guide us into the future, not to dwell in.
For that matter you might want to avoid a GC column with nitrogen in the stationary phase like 1701, 624 (cyanopropyl side chains). Otherwise any non nitrogen solvent is probably fine.

I like
acetone (one of my favorites it dissolves almost anything)
toluene (for higher boilers also has broad solubility)
heptane
diethyl ether (I try to avoid due to risk of peroxide formation)
ethyl acetate
methanol
ethanol
isopropanol
4 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry