Hi
I will be using an FPD detector to analyse samples of petroleum ether for Sulfur species. I read that the FPD gives a quadratic response. I was planning on running different concentrations of sulfur as Carbon Disulfide to see if the method I have is linear. Now I dont know if that would work. Does anyone know if the FPD detector would produce a suitable calibration or will i need to look at using an internal standard as a means of calculating Sulfur content ?
Thanks,
Chemiger
FPDs typically give you the option of taking the square root of the signal automatically, which linearizes the response. The operator's manual or the manufacturer's web site will tell you what you need to know about settings.

Peter
Peter Apps
Thanks Peter will look at the manual today.
The FPD is quite non-linear. You can plot curves but you need multiple levels of standard to accurately develop the curve. The other difficulty is that you may saturate the photomultiplier which will define the upper end of the curve for you.

Likewise, I think you are assuming it is equimolar and it is not. Response is compound specific so you cannot just apply CS2 response factor to other components. Therefore, I am not sure what you are trying to accomplish with the use of an internal standard.

Best regards,

AICMM
4 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry