method validation by GC

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
We initiated Residual solvent method by GC validation. Acetone is one of the solvent. And it is already present in the test sample .. about 1400 ppm. The LOQ established is 50 ppm. How to establish recovery (accuracy) at LOQ....

Regards,
I would NOT perform a 'spike' and 'recovery' experiment at the LOQ concentration because by definition the %RSD at the LOQ is 10.0%. Thus, if you set the acceptance criteria at 2.0 or 5.0% you will fail gloriously! You are demonstrating that your method is consistent and reliable at your expected concentration level. Thus, I would try 50, 100, and 150% of the expected concentrations as long as they are on the linearity curve.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry