Tekmar HT3 Carry over issues

Posted: Mon Jan 14, 2019 7:55 pm
by taylor_chemist
Hello,

I am improving my Headspace -GC - FID method for analysis of terpenes. I am struggling with carry over in between runs. I am using the Tekmar HT3 headspace instrument with the following parameters:

Valve oven temp: 160 C
Transfer line temp: 170 C
Standby flow rate: 50 mL/min
Platen/sample temp: 180 C
Platen temp equil time: 0.00 min
Sample equil time: 0.00 min
Mixer: on
Mixing: Level 5
Mixing time: 10.00 min
Mixer stabilize time: 0.00 min
Pressurize: 9 Psig
Pressurize time: 0.50 min
Pressurize equil time: 0.25 min
Loop fill pressure: 9 Psig
Loop fill time: 0.25 min
Inject time: 1.00 min

I have been injecting a Restek Terpene standard mix #1 usinh methanol as the diluent at a concentration of 120 ppm. I have seen a method which uses this headspace unit using headspace-loop split injection at a split ratio of 10:1 but haven't found this parameter when editing the method.

Please let me know if you have any suggestions or ideas.

Thanks,

Taylor

Re: Tekmar HT3 Carry over issues

Posted: Tue Jan 15, 2019 2:52 pm
by Bigbear
What size is your loop? How do you have the transfer line connected to the GC? I prefer to put them in series with the GC inlet's gas supply. The split ratio should be in the GC's inlet parameters.
You could use a smaller loop instead of splitting.
Carry over can be addressed by either increasing the standby time.

Re: Tekmar HT3 Carry over issues

Posted: Fri Jan 18, 2019 8:53 pm
by taylor_chemist
Hello,

The loop size is 1 mL. The transfer line is connected in parallel with the carrier gas, so I could try to connect them in series.

Thanks so much for your help!

Taylor

Re: Tekmar HT3 Carry over issues

Posted: Fri Jan 18, 2019 9:47 pm
by dlbenach
If what you have written is correct, you are heating your vial hotter than your sample loop or transfer line. Reduce the platen to 150C

Vial temp is 180C - "Platen/sample temp: 180 C" ;
Loop temperature is "Valve oven temp: 160 C";
Transfer line is "Transfer line temp: 170 C"

Dave

Re: Tekmar HT3 Carry over issues

Posted: Fri Jan 18, 2019 11:06 pm
by taylor_chemist
Thanks Dave. I will go ahead and reduce the platen temp to 150 to see if that helps. I appreciate your feedback!

Taylor