Inegration in Chemstation

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
I would like to be able to set a threshold for peak integration for peaks outside my established calibration that doesn't affect the peaks in the calibration. Is this possible in Chemstation?
For example, if I have three peaks I am calibrated for at a peak height of 100 and I want to integrate all unknown peaks present at peak height of 150, is there a way to set the integration to grab only those unknown peaks at 150, but still integrate all three compounds observed in my calibration table regardless of their peak height?
Thanks
Nikko
You may divide the lenght of your run into different time subranges and establish different integration parameters for them if this is what you need.