Analysis of Dual Gas Permeate with GC

Posted: Thu Dec 13, 2018 6:55 pm
by Samuel_Brister
Hello Everyone,

I am constructing a dual gas permeation instrument as my senior engineering design project and will be using GC to determine the permeate gas composition. I am new to this technique, so any information, resources, or suggestions are highly appreciated.

We will be using the instrument to investigate the gas barrier properties of polymer films under competing gases (two of the following; H2, O2, N2, CH4, and CO2) in a dual gas mixture. The permeate gas will be fed to a Hewlett Packard G1540A GC.

The gas lines on the instrument must be made of 316/317 stainless steel because of the hydrogen embrittlement caused by the H2 gas. Should I take this effect into consideration for any stage of the GC process?

I am also having difficulties finding a suitable column for separation of the listed gases. Will I need to use two separate columns and is this possible on the GC I have been provided?

Will I be able to conduct all of the necessary analysis with only the TCD that the GC is equipped with?

Thank you for the help!

Re: Analysis of Dual Gas Permeate with GC

Posted: Sat Dec 15, 2018 3:01 pm
by rb6banjo
Can you provide a drawing of your instrument and sampling apparatus? I’m not really clear on what you mean by “dual” in this context.

It will be tough for you to get all of those analytes with one column. Much discussion has been made on separation of these components. I suggest you search for them using the search box at the top of the main webpage for the forum.

Re: Analysis of Dual Gas Permeate with GC

Posted: Mon Dec 17, 2018 12:07 pm
by pumpedupchemist
You will be able to perform the analysis easily with a dual column but using the PDHID detector.

I can send you an application note if you send me your email address.

Re: Analysis of Dual Gas Permeate with GC

Posted: Mon Dec 17, 2018 4:48 pm
by Samuel_Brister
rb6banjo wrote:
Can you provide a drawing of your instrument and sampling apparatus? I’m not really clear on what you mean by “dual” in this context.

It will be tough for you to get all of those analytes with one column. Much discussion has been made on separation of these components. I suggest you search for them using the search box at the top of the main webpage for the forum.


Here are some schematics of other groups' instruments that I have found in literature. https://imgur.com/a/E0ezHFM
The permeate gas is a mixture of two of the gases I would like to test and I need to perform GC to separate and analyze the permeate gas composition. I have looked at a few of the other posts concerning this topic, but I haven't been able to find any that list a column that I can find and procure.

Thanks!

Re: Analysis of Dual Gas Permeate with GC

Posted: Mon Dec 17, 2018 4:50 pm
by Samuel_Brister
pumpedupchemist wrote:
You will be able to perform the analysis easily with a dual column but using the PDHID detector.

I can send you an application note if you send me your email address.



That would be incredibly helpful! You can contact me at samuel.brister@usm.edu

Thank you!!

Re: Analysis of Dual Gas Permeate with GC

Posted: Tue Dec 18, 2018 1:44 pm
by rb6banjo
Will this do the trick for you?

https://www.restek.com/chromatogram/vie ... C7782-44-7

If the concentrations of the components in your mix are concentrated enough, the TCD should work.

Re: Analysis of Dual Gas Permeate with GC

Posted: Tue Dec 18, 2018 2:24 pm
by Samuel_Brister
rb6banjo wrote:
Will this do the trick for you?

https://www.restek.com/chromatogram/vie ... C7782-44-7

If the concentrations of the components in your mix are concentrated enough, the TCD should work.


That looks great! Thank you for the help!