Overlaying Chromatograms with offset retention times.
Discussions about GC and other "gas phase" separation techniques.
5 posts
Page 1 of 1
Hello. I have an Agilent GC model 7890B. I was wondering if there is any way to overlay chromatograms with offset retention times. I cut the column and reinserted it into the injector and detector. After making sure everything was secure, I started the GC. When I did my first two injections. the retention times were overlaying perfectly. The next day, I notice the retention times are completely off. I am not sure if this means my column is too short to compensate for my sample eluting or I need to install a new column. Is there anything to help with overlaying offset retention times? I use Agilent OpenLab for data analysis.
-
- Posts: 926
- Joined: Mon Feb 13, 2006 6:28 pm
- Location: Sz-n
oh0008 wrote:
... When I did my first two injections. the retention times were overlaying perfectly. The next day, I notice the retention times are completely off. ...
May be a leak in the inlet causing changes in flow of carrier gas through the column resulting in retention time changes.
-
- Posts: 2
- Joined: Mon Nov 14, 2016 9:32 pm
How do I check a leak in the inlet? I do not have a flowmeter to check for leaks.
-
- Posts: 926
- Joined: Mon Feb 13, 2006 6:28 pm
- Location: Sz-n
oh0008 wrote:
How do I check a leak in the inlet? I do not have a flowmeter to check for leaks.
If you don't have leak detector install your column in the inlet again paying attention to do it properly. Change the inlet septum as well if it is "tired" with many injections.
What do you mean by "RT are completely off" ? Are they longer or shorter or they change randomly ?
-
- Posts: 348
- Joined: Mon Sep 13, 2010 8:28 am
About visual comparison:
http://unichrom.com/lnet/c-anone.html
Active chromatogram can be shifted either using shift+mouse dragging or using toolbar "Move graph" buttons.
And yes, the OpenLab .ch file can be opened.
Workbook with two chromatograms can be constructed in two ways:
1. menu Edit/Copy layer, select another workbook and Edit /Paste layer
2. menu Edit/ Collect batch
Visual comparison can help in decision - if it is inlet problem (leak or liner overflow) or artifact from another injection.
http://unichrom.com/lnet/c-anone.html
Active chromatogram can be shifted either using shift+mouse dragging or using toolbar "Move graph" buttons.
And yes, the OpenLab .ch file can be opened.
Workbook with two chromatograms can be constructed in two ways:
1. menu Edit/Copy layer, select another workbook and Edit /Paste layer
2. menu Edit/ Collect batch
Visual comparison can help in decision - if it is inlet problem (leak or liner overflow) or artifact from another injection.
5 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science