Varian Cp-3800 not reaching ready status

Posted: Fri Oct 19, 2018 1:34 pm
by sweezy
Hey guys,

a have a problem concerning the Varian CP-3800 with a 8400 autosampler connected to a MS (but outlet pressure is still atm) and I would be so greatful of you could help me.
I tried to start a new method and here are some key data (if you need more, let me know):
- 20 m column with 0.18 mm ID
- N2 carrier gas
- using pressure pulse for the injection at 35 psi
- oven start temperature: 70 °C
Every parameter is reaching its set point execpt for the column flow. I selected constant column flow and I tried values from 0,4 - 1,0 mL/min and the pressure program is changing all right but the system will still not go to ready. Do you have any idea why this is?

Thanks! :)

Re: Varian Cp-3800 not reaching ready status

Posted: Sat Oct 20, 2018 8:32 am
by antonk
Column flow cannot be measured by instrument, only total flow.
System checks the pressure and evaluates column flow, maybe system still believes the column output is in vacuum?

Re: Varian Cp-3800 not reaching ready status

Posted: Sun Oct 21, 2018 3:11 pm
by sweezy
Thank you for your reply!
I meant the column flow is a set value, depending on which the Varian creates a pressure program. If I change the column flow value, the pressure also changes.
I already changed the vacuum setting, so thats probably not it. Still, thanks a lot!