We are analyzing one compound on Agilent 7890GC A and B with following GC condition.
Column : HP-5, 30m x 0.32mm ID x 0.25 micron film thickness.
Carrier gas : nitrogen (1.5 ml constant flow)
Injector : 250 (Split 100:1)
Detector : 280
Oven : 50 (2) ramp of 100 degree per minute up to 280 hold up to 10 minute.

main peak are eluting at 2.6minute and then baseline got disturbed up to 5 minute.
when we are analyzing same compound with same column on shimadzu 2010/varian 450 GC with all the same parameters we are getting one impurity at 4.3 minute and baseline which we are getting very good.

now I am not able to understand why the baseline disturbed on agilent GC.
Is there any injection technique variation in all the three GC due to that we are not able to get as we had check almost all the inlet parts of agilent GC but still we are not able to get the smooth base line.
IF anyone can help please suggest me where to look to get better baseline.
It would be helpful if you could post an image of what your baseline is doing. In the instructions at the top of this forum, you can read about how to do this.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry