Hi,

I need to analyze glycols with GC-FID, the compounds appear in EPA 8015 and it says that the extraction is by azeotropic distillation with direct injection of water samples. I have a DB-5, is this a column for this type of injection? i think not.
With this column, can i analyze glycols? If not, what other method or column do you recommend?

Thanks.
It has been done using a DB-5 but a wax column works much better. Limit the injection volume with water.
Simple injection of 0.5 uL should get you to low ppm for ethylene and propylene glycol.
So i can use a DB-5 with direct injection of water but with low volume, is that correct?

Thank you very much.
If your detector is an FID you're going to have a hard time keeping that flame lit after injecting water on a "5" column. Wax columns are what we use. We also found we need to be very aggressive with maintenance after injecting water.
Regards,

Christian
I'm betting your glycols won't give you nice peak shapes with the nonpolar DB-5 phase. The wax is your best bet.
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry