Have anyone ever worked with a DANI Master GC?
I think I have a problem with the split module or maybe with the flow module... Someone know something about that?
Thank you!
Discussions about GC and other "gas phase" separation techniques.
Kaliane wrote:
Hello!
Have anyone ever worked with a DANI Master GC?
I think I have a problem with the split module or maybe with the flow module... Someone know something about that?
Thank you!
sana70 wrote:Kaliane wrote:
Hello!
Have anyone ever worked with a DANI Master GC?
I think I have a problem with the split module or maybe with the flow module... Someone know something about that?
Thank you!
Hi
I work with DANI Master GC, what is your problem?
Do you have any error on the GC's Display about flow or pressure?
Kaliane wrote:sana70 wrote:Kaliane wrote:
Hello!
Have anyone ever worked with a DANI Master GC?
I think I have a problem with the split module or maybe with the flow module... Someone know something about that?
Thank you!
Hi
I work with DANI Master GC, what is your problem?
Do you have any error on the GC's Display about flow or pressure?
Hi!
No... There is no error on the GC's display... but the results are not reliable... the area of the peaks are getting smaller as the time is running... For example: in the beggining of the day (9:00), the sample have 124 mg/L... and in 13:30 the same sample have 99 mg/L... the sample analysis came from the same vial...so can't be humam error, for example...
How many is the percentage error in your GC?
Thank you!
sana70 wrote:Kaliane wrote:sana70 wrote:
Hi
I work with DANI Master GC, what is your problem?
Do you have any error on the GC's Display about flow or pressure?
Hi!
No... There is no error on the GC's display... but the results are not reliable... the area of the peaks are getting smaller as the time is running... For example: in the beggining of the day (9:00), the sample have 124 mg/L... and in 13:30 the same sample have 99 mg/L... the sample analysis came from the same vial...so can't be humam error, for example...
How many is the percentage error in your GC?
Thank you!
Hi
Are you sure that the sample have no volatile compound ?
and it was not vaporized, and your injections were repeatable
We have similar problems, please explain how to inject and your sample type and column.
sana70 wrote:
actually we couldn't fix it, we have to report the results by Area%, so when the peak size (all of peaks) were change, it is not affect on Area%
it is because of PFC, sometimes your peak are larger and sometimes smaller because PFC couldn't stable the pressure.
you should calibrate the PFC.
I hope you can fix it.
Forgive my poor English.
Kaliane wrote:sana70 wrote:
actually we couldn't fix it, we have to report the results by Area%, so when the peak size (all of peaks) were change, it is not affect on Area%
it is because of PFC, sometimes your peak are larger and sometimes smaller because PFC couldn't stable the pressure.
you should calibrate the PFC.
I hope you can fix it.
Forgive my poor English.
I don't know what PFC means... Could you help me?