ELUATION OF HYDROGEN IN RGA

Posted: Sun Jun 10, 2018 8:49 am
by MONICAP
Hello,
My name is Mónica and I'm from Portugal.
I've a RGA HiSpeed and I'm having problems with the H2 peak.
It has about 6 times more area that it should.
This happend after I changed de 13X.
It seams that I've 2 peaks eluting together.
Can someone help me with this?
Thank you so much.

Re: ELUATION OF HYDROGEN IN RGA

Posted: Mon Jun 11, 2018 4:40 pm
by muGC
It could be possible that you have Helium present and that Helium and Hydrogen are co-eluting. Their retention times are relatively close, even with a regular (not "Fast") RGA. If that is the case, you need to adjust your column flow to separate the two compounds. Try injecting a standard with helium and no hydrogen to confirm its retention time.

Or your sensitivity is six times as good now and there is no need to worry.

Re: ELUATION OF HYDROGEN IN RGA

Posted: Tue Jun 12, 2018 3:50 pm
by AICMM
This is a micro GC or a full size GC? You are using argon carrier or nitrogen? Your standard has 6X the hydrogen response than it used to have?

More info, better answer...

Best regards,

AICMM

Re: ELUATION OF HYDROGEN IN RGA

Posted: Fri Jun 15, 2018 4:06 pm
by MONICAP
Thank you so much for the help.
I discovered that the problem was in V3. Was damaged. After being replaced the values of the areas returned to normal.
Next time I'll try to be more explicit, I understood that I was not very clear and did not give enough information.
Once more, thank you