I'm certain this has been discussed in the past, but I can't seem to find a post. We have a Varian 3800 with a combipal attached for headspace analysis. We are having problems with our 20 mL vials. The arm is having difficulty returning them to the sample tray. I believe it's because the caps are misaligned. We are using a restek brand of crimp caps. It appears that slight angles on the caps are causing our problems. What's recommended for this system?

Thanks,
Denise
I use P/N: 502 370

from:

www.sun-sri.com on all of my combipal-type samplers (old LEAP, Agilent, EST Analytical). All rely on a magnet for transporting the vials from the tray to the incubator and back, etc. These are nice because they're a combination of aluminum and steel. You can decrimp them to dispose of the contents of the vial before trashing the vial.

I get them through VWR Scientific but it's always a bit of a hassle because they don't appear in the VWR online catalog. We always have to refer back to the old PO number to get the right VWR P/N.
The only explanation I can think of is you are using nonmagnetic aluminum crimp caps or you are not putting the magnet on the injector or your programmed coordinates are off.
MACHEREY-NAGEL #702774 works well for me.
Check that you are not crimping the caps too tight, or holding the crimper at an angle to the vial when you crimp.

Peter