Hello to everyone! I have come up with a tricky task! I am using an SRI 8610C GC system with ECD/FID in a raw and running workplaces atmosphere BTEX samples. The problem is that ALL BTEX peaks are reduced sudenly. One day I make a quality control (QC) 10ppm standard and all its ok,the next day the same standard is ok BUT the 3rd day the same standard is calculated 7 ppm!!! I made new 10ppm standard but again it was measured 6.5-7 ppm!!! Is there a leakage in my system? The liner is already cleaned, the bolts are tightened and the syringe (manual injection) I checked is ok. I have also changed the septum! 30% loss during 3 days??? Is it an FID problem? The column also is brand new, about 2 months old...Any ideas?
Thank you very much
John
I would definitely check that the flame tip of the FID isn't partially blocked.
Removing this and giving it a quick flush with acetone and using a thin piece of wire should clear the debris.

Re-test your sample to see if there is improved sensitivity.

Good Luck!
I re-check the sample after weekend and it is OK!!!! But I will give a clean in the FID wire and collector electrode.....
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: benhutcherson and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry