Method Detection Limit for GC-FID
Discussions about GC and other "gas phase" separation techniques.
12 posts
Page 1 of 1
Hello. I am trying to setup glycol analysis method. I need to know how to determine MDL. What are the steps involved. I have already created the calibration curve. How to determine MDL. Thank you
-
- Posts: 3503
- Joined: Mon Aug 30, 2004 11:43 pm
Keep injecting lower levels until you cannot get required S/N.
I assume that you are developing/validating a new test method, otherwise there's be no need to establish MDL.
On the other hand, my supervisor demanded we establish MDL for an ethanol assay we were validating where the active level was 62%. So MDL was a zillion times lower, so very insignificant. But he was fearful of writing a justification, so we just did it (was a few ppm).
USP monograph for glycerin has test requirements for ethylene glycol/DEG contamination at under 0.1% (which in my opinion is silly, as anyone adulterating glycerin to make money would adulterate at much higher levels).
I assume that you are developing/validating a new test method, otherwise there's be no need to establish MDL.
On the other hand, my supervisor demanded we establish MDL for an ethanol assay we were validating where the active level was 62%. So MDL was a zillion times lower, so very insignificant. But he was fearful of writing a justification, so we just did it (was a few ppm).
USP monograph for glycerin has test requirements for ethylene glycol/DEG contamination at under 0.1% (which in my opinion is silly, as anyone adulterating glycerin to make money would adulterate at much higher levels).
-
- Posts: 56
- Joined: Tue Feb 06, 2018 10:38 pm
Consumer Products Guy wrote:
Keep injecting lower levels until you cannot get required S/N.
I assume that you are developing/validating a new test method, otherwise there's be no need to establish MDL.
On the other hand, my supervisor demanded we establish MDL for an ethanol assay we were validating where the active level was 62%. So MDL was a zillion times lower, so very insignificant. But he was fearful of writing a justification, so we just did it (was a few ppm).
USP monograph for glycerin has test requirements for ethylene glycol/DEG contamination at under 0.1% (which in my opinion is silly, as anyone adulterating glycerin to make money would adulterate at much higher levels).
Lower levels of what. The highest concentration in our samples is 100ppm. So what do i inject to get MDL. How will i know i have the required signal/noise ratio.
-
- Posts: 900
- Joined: Thu Apr 18, 2013 2:10 pm
Lower and lower concentrations of your analyte in your matrix. If it's glycol in water, you need to make standards that are lower and lower in glycol concentration until you get to where your signal is reliably just above the noise. The "best" rule of thumb that most of us use is S/N ratio of 3 for limit-of-detection.
With a mass spec detector, I've had instances where a S/N = 1.75 was good enough for me to be able to tell that what I was seeing was really the analyte. BUT, that was with a very powerful detector. When you're using a flame, the S/N of 3 is probably a good way to go.
With a mass spec detector, I've had instances where a S/N = 1.75 was good enough for me to be able to tell that what I was seeing was really the analyte. BUT, that was with a very powerful detector. When you're using a flame, the S/N of 3 is probably a good way to go.
-
- Posts: 56
- Joined: Tue Feb 06, 2018 10:38 pm
rb6banjo wrote:
Lower and lower concentrations of your analyte in your matrix. If it's glycol in water, you need to make standards that are lower and lower in glycol concentration until you get to where your signal is reliably just above the noise. The "best" rule of thumb that most of us use is S/N ratio of 3 for limit-of-detection.
With a mass spec detector, I've had instances where a S/N = 1.75 was good enough for me to be able to tell that what I was seeing was really the analyte. BUT, that was with a very powerful detector. When you're using a flame, the S/N of 3 is probably a good way to go.
Like how low we are talking for example 0.1 PPM, 0.2 PPM etc
-
- Posts: 478
- Joined: Fri Jun 02, 2017 11:10 pm
- Location: Western Colorado
40 CFR Appendix B to Part 136 - Definition and Procedure for the Determination of the Method Detection Limit-Revision 1.11]40 CFR Appendix B to Part 136 - Definition and Procedure for the Determination of the Method Detection Limit-Revision 1.11 is the official EPA method. Its been updated a few times. I prefered the earlier versions based on the standard deviation of first method blanks and then mdl spiked method blanks. Here is a pdf link Definition and procedure for the Determination of the Method Detection Limit, Revision 2
-
- Posts: 56
- Joined: Tue Feb 06, 2018 10:38 pm
LALman wrote:
40 CFR Appendix B to Part 136 - Definition and Procedure for the Determination of the Method Detection Limit-Revision 1.11]40 CFR Appendix B to Part 136 - Definition and Procedure for the Determination of the Method Detection Limit-Revision 1.11 is the official EPA method. Its been updated a few times. I prefered the earlier versions based on the standard deviation of first method blanks and then mdl spiked method blanks. Here is a pdf link Definition and procedure for the Determination of the Method Detection Limit, Revision 2
OK i will check this out. Thank you
-
- Posts: 104
- Joined: Sun Sep 13, 2015 7:54 am
sohaib1 wrote:
Like how low we are talking for example 0.1 PPM, 0.2 PPM etc
That's the question only you can answer and that's exactly what we're telling you to find out.
Make a solution at 10ppm, can you easily differentiate the signal from the noise? Great! Now make 5ppm and do the same, and test smaller concentrations until YOU find the concentration that you would feel comfortable stating is the MDL based on what you are able to observe and quantify.
-
- Posts: 543
- Joined: Thu May 18, 2006 2:36 pm
- Location: UK
Zoraku has summed up the repeated suggestions of the other posters perfectly!
You can always post chromatograms if you need further help on the S/N as you approach the MDL
You can always post chromatograms if you need further help on the S/N as you approach the MDL
Regards
Ralph
Ralph
-
- Posts: 56
- Joined: Tue Feb 06, 2018 10:38 pm
Zoraku wrote:sohaib1 wrote:
Like how low we are talking for example 0.1 PPM, 0.2 PPM etc
That's the question only you can answer and that's exactly what we're telling you to find out.
Make a solution at 10ppm, can you easily differentiate the signal from the noise? Great! Now make 5ppm and do the same, and test smaller concentrations until YOU find the concentration that you would feel comfortable stating is the MDL based on what you are able to observe and quantify.
OH OK Thank you
-
- Posts: 56
- Joined: Tue Feb 06, 2018 10:38 pm
GOM wrote:
Zoraku has summed up the repeated suggestions of the other posters perfectly!
You can always post chromatograms if you need further help on the S/N as you approach the MDL
Thank you
-
- Posts: 25
- Joined: Wed Feb 28, 2018 4:51 pm
- Location: UK
Hi,
Since you have a calibration you could always use the alternative method based on the ICH's Standard Deviation of the Response and the Slope. The methodology is explained on page 4.24 of the Fusion QbD Method Validation Users Guide http://http://www.ciinformatics.co.uk/?attachment_id=647
The Limit of Quantitation can be calculated similarly (page 4.25). Typically these values are determined during a combined experiment that determines the methods Accuracy, Linearity, and Range. If you are validating the method you will need to determine these as well.
For more information on Fusion QbD Method Validation software start here http://www.ciinformatics.co.uk/products/fusion-method-validation-software/
Kind regards,
Bob
Since you have a calibration you could always use the alternative method based on the ICH's Standard Deviation of the Response and the Slope. The methodology is explained on page 4.24 of the Fusion QbD Method Validation Users Guide http://http://www.ciinformatics.co.uk/?attachment_id=647
The Limit of Quantitation can be calculated similarly (page 4.25). Typically these values are determined during a combined experiment that determines the methods Accuracy, Linearity, and Range. If you are validating the method you will need to determine these as well.
For more information on Fusion QbD Method Validation software start here http://www.ciinformatics.co.uk/products/fusion-method-validation-software/
Kind regards,
Bob
12 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science