MSD leak?

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hi all,

I have a GC7890A system, with a 5975C MSD. Both were serviced in late December and I have since started to notice an increase in the Nitrogen + O2 levels. Before service, levels were below 1% for N2 and ~0.3% for O2. I (stupidly) did not tune after the engineer's visit and did not use the MSD until a few weeks later, when I noticed an increase to 6% in N2 and ~1.5-2% in O2. In all cases, the water has remained below 0.6%. At this point, I also noticed a decrease in the relative abundance of m/z 502 (above 5% to below 3%).

Since my lack of tuning, I can't pinpoint the exact time at which the leak appeared but since the only maintenance on the MSD was the service, I assume it was at that point. I think that the levels of air with low water show that we might have a small leak (?) and if this was the case, I'd rather fix it before it gets worse. I have re-tightened the nut at the interface and did some inlet maintenance in case this was the cause. I am now thinking about checking the o-ring at the side door of the MSD.

1-Do you agree with a small leak being present?
2-If so, are there any other obvious places than the side door to check (we have limited leak-checking capabilities)?
3-Given that we only use our MSD in SIM mode (m/z 62), is the decrease in relative abundance at high mass an issue?

Thanks for help!
I don't think it's much of an issue.

To check if you have a leak at the MS side, use a can of dust remover spray (to remove dirt from your keyboard). Put the MS in scan between 10-100 m/z and spray on the interface nut while checking the mass spectrum. If a peak appears due to spraying, you can tighten it slightly and check again. Same trick for the MS door.
Rndirk wrote:
I don't think it's much of an issue.

To check if you have a leak at the MS side, use a can of dust remover spray (to remove dirt from your keyboard). Put the MS in scan between 10-100 m/z and spray on the interface nut while checking the mass spectrum. If a peak appears due to spraying, you can tighten it slightly and check again. Same trick for the MS door.


This is the best method to find a leak in MS. If you do not see a spike in the spectrum from the dusting spray( normally in the 55-85m/z range depending on the chemical in the spray) then it is most likely just a shift in the relative mass abundance across the scan range due to the maintenance. Since you noticed a lower 502 peak, it could be the 28 and 32 peaks are now a little higher than normal. Another tune may fix it if you don't find a leak.
The past is there to guide us into the future, not to dwell in.
Thanks to both of you for the replies. I will give the spray a go and see what I find.
4 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry