RTL on 7890B coupled with 5977B series MSD
Discussions about GC and other "gas phase" separation techniques.
2 posts
Page 1 of 1
I performed an RTL lock. Locked a 40m column to mannitol. I come back next week to find mannitol isnt eluting at 21.5 anymore. What could be an issue? Leaks?
-
- Posts: 391
- Joined: Tue Dec 30, 2008 2:57 pm
Hi, when do you observe the peak eluting? Earlier or later? What was the original RT and the new RT?
The RT should be quite stable and only need to be relocked after significant column trimming is performed. It is possible you have a leak, if you do an air and water check do you see any problem?
The RT should be quite stable and only need to be relocked after significant column trimming is performed. It is possible you have a leak, if you do an air and water check do you see any problem?
2 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science