Hi all

I am in Australia and have had a HP5890 GC come into my possession from a colleague in the US which is 120V. I need to get a step down transformer to be able to operate it, any ideas on the output in watts i would require?

Many thanks in advance
There should be a panel on the back with the power requirements. Also, if I remember correctly (which is hope over expectation !), you may be able to convert it to 220 V by shifting some jumpers. Do you have manuals ?

Peter
Peter Apps
petehops wrote:
Hi all

I am in Australia and have had a HP5890 GC come into my possession from a colleague in the US which is 120V. I need to get a step down transformer to be able to operate it, any ideas on the output in watts i would require?

Many thanks in advance

This should be very useful:

https://www.agilent.com/cs/library/supp ... a15282.pdf
I know the 5890s required 110v and 20amp circuits if that helps. I believe they pull 2200VA max as listed on the back panel, but has been a while since I saw one.
The past is there to guide us into the future, not to dwell in.
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: benhutcherson and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry