emi_r wrote:Peter Apps wrote:
A shift of a whole chromatogram is actually really unusual, and can only be related to something happening at slightly different times at the beginning of the run. The most likely culprit is that column 2 post run pressure is set to zero post run - when the run starts it has to jump to 240 KPa, which it will never be able to do as repeatably as you want. This is putting a variable back-pressure on column 1 just as the injection is being made, which could well shift the whole chromatogram. How repeatable are your peak areas ?, and does peak area correlate with retention shift ?
Peter
Thank you Peter!
Do you think I can change the pressure to 240 kPa at the end of the run (post run)? Will the pressure be set to 0 after the post run? Or are the settings loaded directly from the next run?
I have now figured out the function of how to record the pressures in each run. I will record them from now on and compare them.
I'm working in constant pressure mode, so all pressures need to be stable during the run, is that right?
I cannot see a reason why you should not set the post run pressure to 240. It is certainly worth a try as a part of troubleshooting if nothing else. You should see the pressure staying at 240 all the time. Yes, pressures and (temperatures) need to be stable or smoothly increasing.
Peter