moisture analysis using GC-TCD
Discussions about GC and other "gas phase" separation techniques.
2 posts
Page 1 of 1
How can I analyse moisture using my GC(TCD detector) with a Molecular Sieve as well as a Poropack column with Helium carrier gas ?
-
- Posts: 1742
- Joined: Fri Feb 09, 2007 10:33 pm
- Location: Albq, NM
You cannot measure moisture with the sieve, it will not elute in any kind of reasonable time. You can probably measure it with the Porapak but the peak shape may be atrocious, depends on which Porapak. Also, you must have a way to inject just on the Porapak to detector, i.e. without the sieve in line.
About the TCD, what kind of detection limits are you hoping to achieve, because with a TCD they will not be that great.
Best regards.
About the TCD, what kind of detection limits are you hoping to achieve, because with a TCD they will not be that great.
Best regards.
2 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science