- Posts: 4
- Joined: Wed Aug 01, 2012 6:57 pm
We have a 6890 GC/MSD set up with 18593B tower/tray and G1512A ALS controller. The ALS is installed on the front column; there is no rear. We run MSD Chemstation B.01.00 on Windows NT4.
The problem is: It appears that the tower has stopped communicating with the ALS controller.
Symptoms began when I went to fire an injection, noticed the bottle never made it to the tower, and there was an "injector communication error".
Upon starting MS Chemstation or trying to load a method, an error dialog appears stating, "The method being loaded requires an injector which is not installed. The method may not be usable. Either correct the configuration, edit the method, or load another method." Trying to fire the injection anyways leads to "Injector is not enabled - check method" error dialog. I can go to Instrument -> Inlet and Injection Parameters and select GC ALS, but the "Injection location" is greyed out.
Manual injection works fine.
I have run the Agilent Injector Diagnostics (document A15704). The tower operates normally; making me suspect a communication problem.
Nothing has changed since the whole set-up had been working earlier that day. The failure happened at about 12:30-1pm. Perhaps it was a power spike or something?
My question: What are some steps I can take to troubleshoot the injector? Is this an issue that I should just call Agilent for?