I used to quantify acrylic acid with an FFAP column,but after 2/3 injections,the performance of the column is bad,tailing etc..
My parameters are on column inj. ffap 30*0.53*1.0
solvent Etoh oven programm70°C 1min-10°C/min-230°C
Does anybody know?
Bye
-------------------------------------------------------------------------------------------------------
By Anonymous on Thursday, June 10, 2004 - 06:05 am:
Polymerization of acrylic acid in the column at high temperature?
-------------------------------------------------------------------------------------------------------
By Anonymous on Thursday, June 10, 2004 - 12:37 pm:
What temperature his high enough for polymerisation?
-------------------------------------------------------------------------------------------------------
By Anonymous on Thursday, June 10, 2004 - 03:22 pm:
Acrylic acid, unless stabilized, will polymerize even at room temperature. At elevated temperatures the rate is much higher. You might want to try HPLC