- admin
- Site Admin
-
- Posts: 647
- Joined: Wed Aug 11, 2004 12:19 am
Dear All:
I am having difficulty in understanding the principle for preparation of standards.
My objective is to develop a calibration curve for ethanol using n-propanol as the internal standard. My range of concentrations for ethanol will be from 80% to 100%.
I usually prepare individual mass fractions of ethanol (i.e., 8g ethanol and 2g of water) and then from this solution I take 1g and add 0.1g of water to dilute it. I use a 4 place analytical balance and weigh accurately the mass. I use a small vial for making the sample prep.
My question is what should be the weight of n-propanol that I should add in each of the varying concentrations of ethanol (i.e, 80,90,95 and100)? Should the weight of n-propanol vary or should it be constant?What should be the concentration of n-propanol that I should add to these ethanol solutions?
I am using a HP 5890 series II with FID and HP-5 column.
Please provide some guidance at the earliest. Thanks in advance.
-------------------------------------------------------------------------------------------------------
By Anonymous on Friday, June 11, 2004 - 08:16 am:
You have only to know what concentration of internal standard is. It does not matter constant it or not. To simplify calculations use constant concentration.
-------------------------------------------------------------------------------------------------------
By Anonymous on Saturday, June 19, 2004 - 04:48 pm:
I run a EtOH method where I use propanol as the ISTD (1N)or (0.031 mL per Liter). You should use a different column. There are columns made for EtOH testing (ALC-1). When making the STDs you should use volume and calulate the wieght of the volume. I think EtOH has a denisty of 0.78 g/mL. The same amount of ISTD should be added to each STD.
-------------------------------------------------------------------------------------------------------
By Anonymous on Monday, June 21, 2004 - 03:53 pm:
Hello:
Thanks for the tip. Can you please tell me how you are preparing the ethanol standards? If u can provide information in detail on this regard it would clear many confusions I have related to this issue.
Thank you
-------------------------------------------------------------------------------------------------------
By Anonymous on Tuesday, June 22, 2004 - 01:38 pm:
For 100 ml solution at 80 % weight/volume: use 80 g of alcohol, 1 ml ISTD and the rest of water. Try a run if you can see the ISTD (propanol), if not increase the amount of ISTD. I dont remember exactly but i belive if you mix ethanol and water that give volumic contraction, verify this.