Good morning all

I am busy validating a residual solvents method. The method will be treated as a limit test. Therefore, the following method validation parameters will be done: Specificity and LOD as per USP <1225>.

This method will be required to be transfered to our QC laboratory. Since this method has been validated as a limit test, what would be required for an AMT?
I was thinking of only requesting the Receing Lab (QC) to confirm the LOD. Would an LOD satisfy AMT requirements?

Thanks in advance.
Michael
If the sample prep is anything more complicated than a dilute and shoot, you're going to want them to repeat some sample preps from lots you've also used, in addition to demonstrating that they get acceptable LOD and suitability results. It's pretty typical intermediate precision stuff, even with limit tests.
Thanks,
DR
Image
LOD is variable depending on the analyst, HPLC, time of day... Just make the LOD acceptance criteria as 'Not More Than (NMT)....
Morning Everyone

Thanks for the suggestions.

Regards
Michael
4 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry