We are having trouble when analyzing the alcohol content (butyl carbitol) of Butyl Carbitol Stearate. When injecting the same sample multiple times the area percent of the alcohol will vary between 1 - 2%. I have checked all the usual suspects: septa, pressure, liner. Does anyone know if the column or other interactions could be accounting for the molecules to be retained somewhere?
-
- Posts: 3510
- Joined: Mon Aug 30, 2004 11:43 pm
We always assayed alcohols and carbitols with more than 3 carbons by using DMF as solvent, derivatizing with BSTFA, and assay on a nonpolar capillary like DB-5.
-
- Posts: 2
- Joined: Wed Nov 15, 2017 3:51 pm
All of our products are run neat. With the exception of a few methylated raw materials (Fatty Acids). I have set up the inlet with a split and we only inject 1ul of the butyl carbitol stearate.
We use the GC as a more accurate measurement of the residual alcohol for in-process samples while the plant is steam stripping the OH out. We went to this method because we don't get accurate numbers from the titration method. But lately we are seeing that duplicate injections are not coming out the same.
We use the GC as a more accurate measurement of the residual alcohol for in-process samples while the plant is steam stripping the OH out. We went to this method because we don't get accurate numbers from the titration method. But lately we are seeing that duplicate injections are not coming out the same.
3 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science