Chromatogram is weird... any ideas?

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi,

We tried to measure MTBE and different alcohols with a GC and instead of serious results we get the exact same periodical chromatogram everytime:

Every 1.5mins there is a very high peak with over 30000 counts. Right after that follows a peak 3000 counts and a peak with approx. 300 counts. Looks like some sort of test run to me...

What could be the reason for that? What could we do?

Thanks for help
~YesPhycisist
The reason is that you have the test signal selected rather than the detector signal.

Gas man
GasMan wrote:
The reason is that you have the test signal selected rather than the detector signal.

Gas man


Yes.