I noticed instability in the total inlet flow, it should have been about 14.6ml/min (the TD uses 10ml/min) but was randomly varying by 0.1-0.2ml/min each way. I disconnected the TD part and it is now a standalone GC again, and the problem is still there. The setup is as follows:
6890 with EPC inlet, column out to Deans splitter at constant 4psi (auxillary EPC), from there to FID and MSD
RT Q-bond PLOT 30m x 320um with particle trap
1.6ml/min constant flow, which is 12psi at 40C (start temp), splitless inlet setting
Total inlet flow 4.6ml/min (1.6 plus 3 for septum purge, unadjustable)
I leak checked the inlet and zeroed the flow and pressure sensors. I tried constant pressure mode, and split (10:1), the former was a little better but not much. I tried this on another 6890 and there was no problem. I also had some pre-move data where I logged flows and this looked fine. I disconnected the column tail from the splitter so it was at a constant pressure (atmospheric) but it made no difference.
It looks like poor EPC control but I don't know if there is anything I've missed or could try. I could try swapping the EPCs from the other 6890 but it is in use so I'd really need to want to. As always I'd very much appreciate your thoughts!
John