we are using a modified cyclodextrin GC column from MN (Hydrodex beta-TBDA) for chiral analysis of an organic acid.
The sample prep is a liquid/liquid extraction to free the acid from its salt using 1M H2SO4 and MTBE. The organic layer is injected.
Here's the problem: The column is damaged after a few dozen injections when we use the extraction sample prep as opposed to the free acid in MTBE. Trimming off a few meters will temporarily restore the separation.
Obviously, we are injecting something that we shouldn't. The column is not compatible with water and the MTBE could contain up to 15 g/L of water and at an injection volume of 1 µL and a split of 40 that's 0.4 ng water on column per injection. Is that enough water to damage a 0.25mm x 1µm df. GC column? Will the presence of H2SO4 worsen the issue.
We are taking measures to either dry the MTBE before injection and/or achieve a better phase separation using the salting out method.
Still, I would be grateful for any input on this issue. many thanks in advance!