GC: Why are some peaks sharp and narrow and some are wide?

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Dear everyone,

Why do some analytes show sharp and narrow peaks and some analytes show broad peaks in gas chromatography?
Attached is a picture of a chromatogram of various organic solvents where you can see how methoxy-2-propanol and butanol show broad peaks compared to other analytes. Regardless of the concentration, these two analytes always appear on the chromatogram as a broad peak.
Does the temperature program of the oven have an effect on the appearance of peaks?

https://ibb.co/LQZJyjw
I get the same thing with my Glycols analysis and would like to know why.
https://photos.app.goo.gl/soyBceYcpsisxDdz6

Image

The above chromatogram has a broad peak for ethylene glycol at 6.2 mins and a nice sharp peak for diethylene glycol at 7.8 mins.

The matrix is water and the method is :

Sample Volume (uL): 0.5
Mode Pulsed Splitless
Heater On 280 °C
Pressure On 3.2572 psi
Total Flow On 64.5 mL/min
Septum Purge Flow On 3 mL/min
Gas Saver On 20 After 2 min mL/min
Injection Pulse Pressure 20 psi Until 0.5 min
Purge Flow to Split Vent 60 mL/min at 1 min
Liner AG0-8173 Liner for Agilent (4 mm ID x 78.5 mm L x 6.3 mm OD), Split/Splitless, Dual Taper, 5/Pk
Oven
Equilibration Time 1 min
Max Temperature 250 °C
Maximum Temperature Override Disabled
Slow Fan Disabled
Temperature
Setpoint On
(Initial) 40 °C
Hold Time 0 min
Post Run 240 °C
Program
#1 Rate 20 °C/min
#1 Value 210 °C
#1 Hold Time 0 min
Column Information Phenomenex 7HG-G013-11: 1135458
ZB-WAXplus
Temperature Range -60 °C—250 °C (250 °C)
Dimensions 30 m x 250 μm x 0.25 μm
Column lock Unlocked
In Front SS Inlet H2
Out MSD
(Initial) 40 °C
Pressure 3.2572 psi
Flow 1.5 mL/min
Thermal Aux 2 (MSD Transfer Line)
Temperature
Setpoint On
(Initial) 250 °C
Post Run 0 °C
[SIM Parameters]
Group 1 Group ID : Glycols (227)
Resolution : 1
Group Start Time : 5
Number of Ions : 6
Ions
Dwell In Group :( Mass, Dwell) ( Mass, Dwell) ( Mass, Dwell)
( 31.10,50 ) ( 33.10,50 ) ( 45.10,50 )
( 62.00,50 ) ( 75.00,50 ) ( 76.00,50 )
Usually that would be consentration overloading, but you have a polar column which should handle the components you have.

It might be a solvent effect from the water temporarily affecting the polarity of a short section of the column. It might improve if you start the temperature programme at a higher temperature and then ramp more slowly.
Peter Apps
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry