- Posts: 6
- Joined: Wed Jan 04, 2017 12:11 am
The GC is not quite a year in operation, but in the last couple of months the FID output signal has started to fade. Up until lately the FID baseline has been around 10pA, now it's around 1.1. During a column back flush period, when no sample is going to the FID, the signal drops below 1.0 and the unit thinks the FID flame is out and tries to relight. The only other symptom is ethanol QC checks are coming in low on a fresh cal, (88-92% recovery). I think I may have a carrier leak? or flows aren't what the PCM think they are?
When the unit is not in service I manually shut down the flame but I keep N2 flow going and temps at normal. Good or bad?
The only config change is I replaced the Agilent block style gas filter system with in-line filters. The block filters were too short lived and they always leaked. Other than that it's been regular PM, liner, oring, gold-seal, and periodically I clean the FID jet.
Any suggestions?