Here's the rundown:
The N2 level is about 10% at a column flow rate of 1 ml/min. We are using ultra pure He as carrier. The N2 percentage INCREASES to about 30-40% when you increase the column flow from 1 ml/min to 2 ml/min.
The mass spec is not leaking. When you plug off the transfer line, there is no air/water present.
Everything in the oven is tight, using the new hand-tightened inlet/interface nuts from Agilent. All parts are Agilent including the column.
Liner/septum is installed correctly.
The entire flow path has been replaced, by Agilent. So EPC, inlet, split vent line, etc. No change. The ENTIRE GC was then replaced by Agilent. No change.
The filter installed with the system went bad within a week (saturated with O2, indicator turned completely black). Not surprising given there's air in the system. We assumed this meant out gas was bad. However, the entire laboratory is fed on a 4-tank system. No other GC/MS systems show any nitrogen leak.
The GC/MS are set up on benches in pairs. So the gas line splits 3 times after the tanks to feed 3 separate benches. We assumed this particular bench had a leak (previously not in use). We hooked the new GC up to one of the lines feeding one of our other GC/MS. No change.
So then we hooked up one of our GC/MS already in use to the gas line on the bench for the new GC/MS. No leak shows on the older system. Gas line is fine. To be extra sure of this, we rolled a fresh, new tank out and hooked it up directly to the EPC of the new system. Still leaking. This occurs with and without a filter/scrubber on the gas line. With a scrubber, the O2 amount is decreased (this should imply the air is entering the line BEFORE the scrubber, but we find no leak in the gas line).
Agilent performed a leak test on the inlet. Passed to their specifications.
We've changed the column, twice. As it turns out, the new finger tightening nuts use the same ferrules for inlet and interface, so we had the wrong ferrule in the interface nut. Fixed this. No change.
I'm at the end of my rope. Agilent is at the end of theirs. Any help or insight would be greatly appreciated. I have essentially taken this thing apart and put it back together and I still cannot find anything. The fact that the air leak increases with flow would imply it's somewhere in the flow path, but we've replaced that. Twice. And nothing.